b Switch ON Test. The Switch ON Test checks that power is correctly supplied to the
electronics pack and internal diagnostics discover no system fault. The test is conducted as
follows:
● Extend the lower shaft at least 150mm (6ins).
● Hold the coil at least 600mm (24ins) from the ground and away from any metallic
objects and switch on.
● Ensure that four rising tones occur over approximately 12 seconds. A steady
threshold tone should be heard a few seconds after the rising tones finish.
NOTE
On occasion, it is possible that more than four rising tones occur. This does not represent
a fault provided the rising tones finish and a steady threshold tone is produced.
Figure 3: Switch ON Test
● If the correct tones are not heard, check that good batteries are being used and that
the batteries are fitted correctly.
● If the correct tones are still not heard refer to Section 4 for faultfinding procedures.
c. Noise Cancel Test. The test ensures the F3 successfully completes the noise
cancel procedure required whenever local electrical interference is present. During the test,
the Noise Cancel button, wiring loom and electronics pack are checked for correct operation.
The test is conducted as follows:
● Hold the coil stationary and at least 600mm (24ins) above the ground and away from
any large metal objects during the test.
● Press and
immediately release
the Noise Cancel button. (Figure 4)
Figure 4: Noise Cancel
NOTE
The coil should not be moved nor should metallic objects be brought near the coil during this
procedure.
● If functioning correctly, Noise Cancel will commence with two single beeps followed
by 45 seconds of sharp double beeps and finishing with four single beeps.